2016-06-10 — S 5939

Calibration certificate details for 2016-06-10.

Certificate Metadata

Certificate Number
S 5939
Calibration Date
2016-06-10
Due Date
Manufacturer
DH Instruments Inc.
Model
molbox 1
Serial Number
880
Description
Elektrisches Druckmessgerät / Widerstände für Durchflussmessung
Customer
Analyt-MTC GmbH, Klosterrunsstraße 18, DE-79379 Müllheim
Issuing Lab
Europascal GmbH
Adjusted
No
Remarks
Pressure calibration: Es wurde keine Justage durchgeführt (No adjustment was performed). PA (upstream) = 7 Pa / PM = 0,99940; PA (downstream) = 18,3 Pa / PM = 0,99999. Resistance calibration: Es wurde keine Justage durchgeführt (Adjustment has not been performed). Calibration performed according to DKD-R 6-1 (Edition 03/2014), Procedure A for pressure, and EURAMET cg-15_v_3.0 Guidelines for resistance. Pressure reference standard: Kolbenmanometer PG 7601, DH Instruments Inc., PTB 30283/11, valid June 2011 to June 2016. Resistance reference standards: DHI 100 Ohm Simulator SN 121 (cert 11832) and DHI 110 Ohm Simulator SN 125 (cert 11833). Environmental conditions - Pressure calibration: Temperature (22.3 ± 0.5) °C, Ambient pressure (999-1000 ± 2) mbar, Location D-63456 Hanau, g_lokal = 9.81050 94 m/s². Resistance calibration: Temperature (24.0 ± 1.0) °C, Relative humidity (30 ± 15) %, Location D-63456 Hanau. Device has both A and B channels despite being labeled molbox 1 (typically single-channel). Coverage factor k=2 for all measurements.
Certificate PDF
Download

Pressure Upstream (mbar)

Upstream (HI) pressure transducer calibration with hysteresis testing

NominalReference (As Found)DUT (As Found)UncertaintyDev/Spec (As Found)
350350.140349.7590.044-0.68
350350.140349.7440.028-0.71
550551.087550.8090.030-0.50
550551.087550.7770.030-0.55
11001105.9501105.8200.038-0.23
11001105.9501105.7930.038-0.28
16501658.6871658.5990.048-0.16
16501658.6871658.5850.047-0.18
22002207.1782207.0490.058-0.23
22002207.1782207.0550.058-0.22
27502759.9132759.7400.069-0.28
27502759.9132759.7600.068-0.24
33003314.7753314.5740.080-0.30
33003314.7753314.5980.079-0.26
38253867.1393866.8920.091-0.34
38253867.1393866.9110.091-0.31
43754419.054418.760.10-0.37
43754419.054418.780.10-0.34
49254973.074972.770.11-0.36
49254973.074972.770.11-0.36
54755523.215522.890.13-0.36
54755523.215522.860.13-0.39

Pressure Downstream (mbar)

Downstream (LO) pressure transducer calibration with hysteresis testing

NominalReference (As Found)DUT (As Found)UncertaintyDev/Spec (As Found)
350350.140349.8810.037-0.46
350350.140349.8850.022-0.46
550551.087550.8810.026-0.37
550551.087550.8830.024-0.36
11001105.9501105.8350.035-0.21
11001105.9501105.8460.034-0.19
16501658.6871658.6120.046-0.13
16501658.6871658.6210.044-0.12
22002207.1782207.0810.056-0.17
22002207.1782207.0990.056-0.14
27502759.9132759.7860.068-0.20
27502759.9132759.8150.066-0.16
33003314.7753314.6360.078-0.20
33003314.7753314.6670.078-0.16
38253867.1393866.9670.089-0.23
38253867.1393866.9940.090-0.20
43754419.054418.840.10-0.27
43754419.054418.860.10-0.24
49254973.074972.860.11-0.25
49254973.074972.870.11-0.24
54755523.215522.980.12-0.26
54755523.215522.950.12-0.29

Resistance (Ω)

RTD resistance measurement verification for flow measurement

NominalReference (As Found)DUT (As Found)UncertaintyDev/Spec (As Found)
10099.989299.9960.004+0.17
10099.992999.9930.004+0.00
10099.989299.9950.004+0.15
10099.992999.9930.004+0.00
110109.9865109.9970.004+0.26
110109.9845109.9910.004+0.16
110109.9865109.9990.004+0.31
110109.9845109.9940.004+0.24