Power-on sequence and self-test verification for the molbox
terminal.
About this task
Perform this procedure at the start of each measurement session to verify that
the molbox system is operating correctly before making flow measurements.
Procedure
-
Push the power switch on the rear panel.
The molbox initializes, performs error checks, calibrates its
internal ohmic measurement system, and displays the main run screen. The
top left of the screen should show a flow value near zero or
<BPR HI>.
-
Verify that the main run screen does not display
<NO BLOC>.
If <NO BLOC> is shown, the molbox has not
detected a molbloc connection. Check that the molbloc electrical cable
is properly connected and push
SETUP, <1molbloc> to
reload the molbloc data.
-
Check the pressure measurement by connecting the molbloc to the molbox,
shutting off the gas supply, and opening both molbloc ends to atmospheric
pressure.
-
Push P&T to display the pressure and
temperature screen.
-
Observe the upstream and downstream absolute pressures.
-
Verify both readings indicate current atmospheric pressure and
agree within ±0.02 % of the nominal full-scale calibrated
range.
Note: The pressure values on the P&T screen are corrected for TARE and
AUTOZERO. If the readings disagree by more than the tolerance, one or
both RPTs may need calibration or service.
-
Check the temperature measurement.
-
From the P&T screen, push P&T
again to display the temperature screen.
-
Verify that both molbloc PRT readings agree within ±0.2 °C and
approximately match ambient temperature.
Allow 30 to 60 minutes of warm-up in a stable temperature environment
before evaluating temperature agreement. If the two readings disagree
by more than ±0.2 °C, there may be a problem with the molbloc or the
molbox TEMPERATURE MEASUREMENT function.
-
Push ESCAPE to return to the main run
screen.
-
Perform a TARE at the operating line pressure before making
measurements.
Push TARE to zero the pressure transducers at
the current line pressure. This removes offset errors from the pressure
readings.
Results
The molbox is ready for flow measurements when all verification checks pass and
the main run screen displays a flow value with the <*>
Ready indicator.
CAUTION: Set the security level to low with password protection
to prevent inadvertent changes to critical internal settings. See the Fluke
manual section on security levels.