Multimeter BK5491A — Drift Analysis

Drift analysis and trend monitoring for Multimeter BK5491A.

Assessment

TODO: Review drift history and write metrological assessment.

Internal Specification: Analyt MTC — BK5491A 1-year accuracy specifications

Manufacturer Specification: BK Precision 5491A Manual — 1-year accuracy specifications

DC Current (A)

Setpoint (A)20242025
As-found Dev/Spec — measured error normalized to internal specification
0+0.40+0.40
0.0004+0.40+0.40
0.004+0.38+0.33
0.04+0.00+0.04
0.4+0.08+0.04
2-0.49-0.39
Inter-calibration drift Dev/Spec (Δ) — change since previous calibration, normalized
0+0.00
0.0004+0.00
0.004-0.04
0.04+0.04
0.4-0.04
2+0.10

Figure 1 shows each calibration's as-found deviation normalized to the internal specification. Values at ±1.0 reach the specification limit. The shaded blue envelope covers the full spread across all setpoints; the amber band adds measurement uncertainty (RSS). For the full methodology, see Equipment Monitoring & Drift Analysis.

Figure 1. DC Current Drift Trend

Figure 2 shows the drift between consecutive calibrations — how much the device changed since it was last calibrated or adjusted. This isolates the device’s instability from corrections applied during calibration.

Figure 2. DC Current — Drift Between Calibrations

DC Voltage (V)

Setpoint (V)20242025
As-found Dev/Spec — measured error normalized to internal specification
-490-0.07-0.22
-49-0.14-0.36
-4.9-0.22-0.51
-2.5+0.00-0.33
-1+0.33-0.17
-0.5+0.40+0.20
0+0.05+0.03
0.01+0.00+0.00
0.25+0.08+0.08
0.49+0.07+0.12
0.5+0.60+0.20
1+0.50+0.33
2.5+0.56+0.56
4.9+0.58+0.65
49+0.22+0.36
490+0.14+0.29
750+0.00+0.18
Inter-calibration drift Dev/Spec (Δ) — change since previous calibration, normalized
-490-0.14
-49-0.22
-4.9-0.29
-2.5-0.33
-1-0.50
-0.5-0.20
0-0.03
0.01+0.00
0.25+0.00
0.49+0.05
0.5-0.40
1-0.17
2.5+0.00
4.9+0.07
49+0.14
490+0.14
750+0.18

Figure 3 shows each calibration's as-found deviation normalized to the internal specification. Values at ±1.0 reach the specification limit. The shaded blue envelope covers the full spread across all setpoints; the amber band adds measurement uncertainty (RSS). For the full methodology, see Equipment Monitoring & Drift Analysis.

Figure 3. DC Voltage Drift Trend

Figure 4 shows the drift between consecutive calibrations — how much the device changed since it was last calibrated or adjusted. This isolates the device’s instability from corrections applied during calibration.

Figure 4. DC Voltage — Drift Between Calibrations