SMU ADT220 — Drift Analysis

Drift analysis and trend monitoring for SMU ADT220.

Overview

Status: Fit for continued use.

Internal Specification
Analyt MTC — Additel 220 1-year accuracy specifications
Manufacturer Specification
Additel 220 Datasheet — 1-year accuracy specifications
Calibration Provider(s)
EuroPascal GmbH (2019–2025)
Calibration Cycles
6 cycles (2019–2025)

This is a multifunction loop calibrator used in the calibration workflow to supply analog setpoint signals (0–5 V or 4–20 mA) to devices under test and read their analog measurement output. All calibrations have been performed by EuroPascal GmbH. The device has six calibration cycles covering five functions: DC voltage measure, DC voltage source, DC current measure, DC current source, and 24V loop supply output.

Context. The device has never been adjusted — all six calibrations returned as-found values within specification across all functions. Most setpoints have low signal-to-noise ratio (SNR < 3), particularly the source functions where the specification is tight relative to calibration uncertainty.

DC 24V Output (V)

Setpoint (V)20192020202220242025Trend
As-found Dev/Spec — measured error normalized to internal specification
24-0.08-0.2-0.2-0.2-0.2
Inter-calibration drift Dev/Spec (Δ) — change since previous calibration, normalized
24-0.12000

Figure 1 shows each calibration's as-found deviation normalized to the internal specification. Values at ±1.0 reach the specification limit. The shaded blue envelope covers the full spread across all setpoints; the amber band adds measurement uncertainty. For the full methodology, see Equipment Monitoring & Drift Analysis.

Figure 1. DC 24V Output Drift Trend

Figure 2 shows the drift between consecutive calibrations — how much the device changed since it was last calibrated or adjusted. This isolates the device’s instability from corrections applied during calibration.

Figure 2. DC 24V Output — Drift Between Calibrations

The 24V loop supply has been measured in five of six cycles (not in 2021). The Dev/Spec settled to −0.40 from 2020 onward and has not changed since — consistent with a fixed offset within the output's ±0.5 V specification. No concern.

DC Current Measure (mA)

Setpoint (mA)201920202021202220242025Trend
As-found Dev/Spec — measured error normalized to internal specification
-290.00060.0009-0.00040.0001-0.0006-0.0001
4-0.0001-0.0001000.00010
12-0.0002-0.00030.00020.00010.00030
20-0.0004-0.00050.00040.00010.00050.0001
29-0.0007-0.00090.0004-0.00010.00050
Inter-calibration drift Dev/Spec (Δ) — change since previous calibration, normalized
-290.0003-0.00130.0005-0.00070.0005
400.000100.0001-0.0001
12-0.00010.0005-0.00010.0002-0.0003
20-0.00010.0009-0.00030.0004-0.0004
29-0.00020.0013-0.00050.0006-0.0005

Figure 3 shows each calibration's as-found deviation normalized to the internal specification. Values at ±1.0 reach the specification limit. The shaded blue envelope covers the full spread across all setpoints; the amber band adds measurement uncertainty. For the full methodology, see Equipment Monitoring & Drift Analysis.

Figure 3. DC Current Measure Drift Trend

Figure 4 shows the drift between consecutive calibrations — how much the device changed since it was last calibrated or adjusted. This isolates the device’s instability from corrections applied during calibration.

Figure 4. DC Current Measure — Drift Between Calibrations

All five setpoints are within specification across six cycles. The scatter is small — worst case |Dev/Spec| of 0.20 (−29 mA, 2020). All trend forecasts indicate stable behaviour.

DC Current Source (mA)

Setpoint (mA)201920202021202220242025Trend
As-found Dev/Spec — measured error normalized to internal specification
40.00050.00060.00060.00060.00060.0005
80.00040.00040.00040.00050.00040.0002
120.00020.00010.00020.00030.00030
160.00030.000300.0002-0.0007-0.0002
200.00040.00040.00020.0004-0.0007-0.0001
Inter-calibration drift Dev/Spec (Δ) — change since previous calibration, normalized
40.0001000-0.0001
8000.0001-0.0001-0.0002
12-0.00010.00010.00010-0.0003
160-0.00030.0002-0.00090.0005
200-0.00020.0002-0.00110.0006

Figure 5 shows each calibration's as-found deviation normalized to the internal specification. Values at ±1.0 reach the specification limit. The shaded blue envelope covers the full spread across all setpoints; the amber band adds measurement uncertainty. For the full methodology, see Equipment Monitoring & Drift Analysis.

Figure 5. DC Current Source Drift Trend

Figure 6 shows the drift between consecutive calibrations — how much the device changed since it was last calibrated or adjusted. This isolates the device’s instability from corrections applied during calibration.

Figure 6. DC Current Source — Drift Between Calibrations

All five setpoints are within specification across six cycles. The 4 mA setpoint shows a consistent positive offset (~0.26 to 0.32) — this is the lowest setpoint where the fixed offset component dominates the specification, not a drift trend. All trend forecasts indicate stable behaviour.

DC Voltage Measure (V)

Setpoint (V)201920202021202220242025Trend
As-found Dev/Spec — measured error normalized to internal specification
-550.00080.0006-0.0031-0.0006-0.0021-0.0017
-0.29-7e-06-7e-06-2e-05-1.8e-05-2.1e-05-1.4e-05
0.01000-3e-0600
0.051e-062e-062e-0603e-061e-06
0.13e-063e-065e-063e-066e-064e-06
0.298e-068e-061.8e-051.3e-052e-051.3e-05
10000.000100
15-0.0001000.00040.00080.0005
2500.00010.00170.00080.00140.0011
55-0.000300.00340.00150.00290.0022
Inter-calibration drift Dev/Spec (Δ) — change since previous calibration, normalized
-55-0.0002-0.00370.0025-0.00150.0004
-0.290-1.3e-052e-06-3e-067e-06
0.0100-3e-063e-060
0.051e-060-2e-063e-06-2e-06
0.102e-06-2e-063e-06-2e-06
0.2901e-05-5e-067e-06-7e-06
1000.0001-0.00010
150.000100.00040.0004-0.0003
250.00010.0016-0.00090.0006-0.0003
550.00030.0034-0.00190.0014-0.0007

Figure 7 shows each calibration's as-found deviation normalized to the internal specification. Values at ±1.0 reach the specification limit. The shaded blue envelope covers the full spread across all setpoints; the amber band adds measurement uncertainty. For the full methodology, see Equipment Monitoring & Drift Analysis.

Figure 7. DC Voltage Measure Drift Trend

Figure 8 shows the drift between consecutive calibrations — how much the device changed since it was last calibrated or adjusted. This isolates the device’s instability from corrections applied during calibration.

Figure 8. DC Voltage Measure — Drift Between Calibrations

All ten setpoints are within specification across six cycles, spanning two ranges (300 mV and 60 V). The mV-range setpoints near zero are very quiet. The higher-voltage setpoints show some scatter — worst case −0.48 (−0.29 V, 2024) — but no systematic trend. All trend forecasts indicate stable behaviour.

DC Voltage Source (V)

Setpoint (V)201920202021202220242025Trend
As-found Dev/Spec — measured error normalized to internal specification
-0.0091e-061e-061e-06-3e-06-1e-061e-06
0.012e-061e-062e-06-1e-061e-061e-06
0.0753e-062e-062e-0602e-061e-06
0.1352e-061e-061e-0602e-060
0.192e-060002e-06-1e-06
20.00010.00010.00010.00020.00020.0001
40.00040.00040.00050.00050.00050.0004
60.00040.00050.00050.00060.00060.0005
80.00070.00080.00090.0010.0010.0008
100.00070.00080.0010.00110.00110.0008
Inter-calibration drift Dev/Spec (Δ) — change since previous calibration, normalized
-0.00900-4e-062e-062e-06
0.01-1e-061e-06-3e-062e-060
0.075-1e-060-2e-062e-06-1e-06
0.135-1e-060-1e-062e-06-2e-06
0.19-2e-06002e-06-3e-06
2000.00010-0.0001
400.000100-0.0001
60.000100.00010-0.0001
80.00010.00010.00010-0.0002
100.00010.00020.00010-0.0003

Figure 9 shows each calibration's as-found deviation normalized to the internal specification. Values at ±1.0 reach the specification limit. The shaded blue envelope covers the full spread across all setpoints; the amber band adds measurement uncertainty. For the full methodology, see Equipment Monitoring & Drift Analysis.

Figure 9. DC Voltage Source Drift Trend

Figure 10 shows the drift between consecutive calibrations — how much the device changed since it was last calibrated or adjusted. This isolates the device’s instability from corrections applied during calibration.

Figure 10. DC Voltage Source — Drift Between Calibrations

All ten setpoints are within specification across six cycles, spanning two ranges (200 mV and 12 V). The mV-range setpoints are very quiet. The higher-voltage setpoints (4–10 V) show a mild positive drift from 2019 to 2022, peaking at 0.45 (8 V, 2022), then corrected slightly in 2025. All trend forecasts indicate stable behaviour.

Conclusion

Calibration interval. The interval is extended from two years to four years. Electrical measurement uncertainty is negligible compared to mass flow calibration uncertainty — the dominant contributors are the flow reference standards, not the electrical signal chain. The device has never required adjustment across six cycles, and all functions are comfortably within specification. No action required.