SMU ADT220 — Drift Analysis

Drift analysis and trend monitoring for SMU ADT220.

Assessment

TODO: Review drift history and write metrological assessment.

Internal Specification: Analyt MTC — Additel 220 1-year accuracy specifications

Manufacturer Specification: Additel 220 Datasheet — 1-year accuracy specifications

DC 24V Output (V)

Setpoint (V)20192020202220242025
As-found Dev/Spec — measured error normalized to internal specification
24-0.16-0.40-0.40-0.40-0.40
Inter-calibration drift Dev/Spec (Δ) — change since previous calibration, normalized
24-0.24+0.00+0.00+0.00

Figure 1 shows each calibration's as-found deviation normalized to the internal specification. Values at ±1.0 reach the specification limit. The shaded blue envelope covers the full spread across all setpoints; the amber band adds measurement uncertainty (RSS). For the full methodology, see Equipment Monitoring & Drift Analysis.

Figure 1. DC 24V Output Drift Trend

Figure 2 shows the drift between consecutive calibrations — how much the device changed since it was last calibrated or adjusted. This isolates the device’s instability from corrections applied during calibration.

Figure 2. DC 24V Output — Drift Between Calibrations

DC Current Measure (mA)

Setpoint (mA)201920202021202220242025
As-found Dev/Spec — measured error normalized to internal specification
-29+0.14+0.20-0.09+0.02-0.14-0.02
4-0.05-0.05+0.00+0.00+0.05+0.00
12-0.07-0.11+0.07+0.04+0.11+0.00
20-0.11-0.14+0.11+0.03+0.14+0.03
29-0.16-0.20+0.09-0.02+0.11+0.00
Inter-calibration drift Dev/Spec (Δ) — change since previous calibration, normalized
-29+0.07-0.30+0.11-0.16+0.11
4+0.00+0.05+0.00+0.05-0.05
12-0.04+0.19-0.04+0.07-0.11
20-0.03+0.26-0.09+0.11-0.11
29-0.05+0.30-0.11+0.14-0.11

Figure 3 shows each calibration's as-found deviation normalized to the internal specification. Values at ±1.0 reach the specification limit. The shaded blue envelope covers the full spread across all setpoints; the amber band adds measurement uncertainty (RSS). For the full methodology, see Equipment Monitoring & Drift Analysis.

Figure 3. DC Current Measure Drift Trend

Figure 4 shows the drift between consecutive calibrations — how much the device changed since it was last calibrated or adjusted. This isolates the device’s instability from corrections applied during calibration.

Figure 4. DC Current Measure — Drift Between Calibrations

DC Current Source (mA)

Setpoint (mA)201920202021202220242025
As-found Dev/Spec — measured error normalized to internal specification
4+0.26+0.32+0.32+0.32+0.32+0.26
8+0.15+0.15+0.15+0.19+0.15+0.07
12+0.06+0.03+0.06+0.09+0.09+0.00
16+0.07+0.07+0.00+0.05-0.16-0.05
20+0.08+0.08+0.04+0.08-0.14-0.02
Inter-calibration drift Dev/Spec (Δ) — change since previous calibration, normalized
4+0.05+0.00+0.00+0.00-0.05
8+0.00+0.00+0.04-0.04-0.07
12-0.03+0.03+0.03+0.00-0.09
16+0.00-0.07+0.05-0.21+0.12
20+0.00-0.04+0.04-0.22+0.12

Figure 5 shows each calibration's as-found deviation normalized to the internal specification. Values at ±1.0 reach the specification limit. The shaded blue envelope covers the full spread across all setpoints; the amber band adds measurement uncertainty (RSS). For the full methodology, see Equipment Monitoring & Drift Analysis.

Figure 5. DC Current Source Drift Trend

Figure 6 shows the drift between consecutive calibrations — how much the device changed since it was last calibrated or adjusted. This isolates the device’s instability from corrections applied during calibration.

Figure 6. DC Current Source — Drift Between Calibrations

DC Voltage Measure (V)

Setpoint (V)201920202021202220242025
As-found Dev/Spec — measured error normalized to internal specification
-55+0.09+0.07-0.36-0.07-0.25-0.20
-0.29-0.16-0.16-0.45-0.41-0.48-0.32
0.01+0.00+0.00+0.00-0.19+0.00+0.00
0.05+0.05+0.10+0.10+0.00+0.15+0.05
0.1+0.12+0.12+0.20+0.12+0.24+0.16
0.29+0.18+0.18+0.41+0.30+0.45+0.30
1+0.00+0.00+0.00+0.03+0.00+0.00
15-0.02+0.00+0.00+0.09+0.18+0.11
25+0.00+0.02+0.31+0.15+0.25+0.20
55-0.04+0.00+0.40+0.18+0.34+0.26
Inter-calibration drift Dev/Spec (Δ) — change since previous calibration, normalized
-55-0.02-0.44+0.29-0.18+0.05
-0.29+0.00-0.30+0.05-0.07+0.16
0.01+0.00+0.00-0.19+0.19+0.00
0.05+0.05+0.00-0.10+0.15-0.10
0.1+0.00+0.08-0.08+0.12-0.08
0.29+0.00+0.23-0.11+0.16-0.16
1+0.00+0.00+0.03-0.03+0.00
15+0.02+0.00+0.09+0.09-0.07
25+0.02+0.29-0.16+0.11-0.05
55+0.04+0.40-0.22+0.16-0.08

Figure 7 shows each calibration's as-found deviation normalized to the internal specification. Values at ±1.0 reach the specification limit. The shaded blue envelope covers the full spread across all setpoints; the amber band adds measurement uncertainty (RSS). For the full methodology, see Equipment Monitoring & Drift Analysis.

Figure 7. DC Voltage Measure Drift Trend

Figure 8 shows the drift between consecutive calibrations — how much the device changed since it was last calibrated or adjusted. This isolates the device’s instability from corrections applied during calibration.

Figure 8. DC Voltage Measure — Drift Between Calibrations

DC Voltage Source (V)

Setpoint (V)201920202021202220242025
As-found Dev/Spec — measured error normalized to internal specification
-0.009+0.08+0.08+0.08-0.24-0.08+0.08
0.01+0.16+0.08+0.16-0.08+0.08+0.08
0.075+0.12+0.08+0.08+0.00+0.08+0.04
0.135+0.05+0.03+0.03+0.00+0.05+0.00
0.19+0.04+0.00+0.00+0.00+0.04-0.02
2+0.10+0.10+0.10+0.20+0.20+0.10
4+0.29+0.29+0.36+0.36+0.36+0.29
6+0.22+0.28+0.28+0.33+0.33+0.28
8+0.32+0.36+0.41+0.45+0.45+0.36
10+0.27+0.31+0.38+0.42+0.42+0.31
Inter-calibration drift Dev/Spec (Δ) — change since previous calibration, normalized
-0.009+0.00+0.00-0.33+0.16+0.16
0.01-0.08+0.08-0.24+0.16+0.00
0.075-0.04+0.00-0.08+0.08-0.04
0.135-0.03+0.00-0.03+0.05-0.05
0.19-0.04+0.00+0.00+0.04-0.06
2+0.00+0.00+0.10+0.00-0.10
4+0.00+0.07+0.00+0.00-0.07
6+0.06+0.00+0.06+0.00-0.06
8+0.05+0.05+0.05+0.00-0.09
10+0.04+0.08+0.04+0.00-0.12

Figure 9 shows each calibration's as-found deviation normalized to the internal specification. Values at ±1.0 reach the specification limit. The shaded blue envelope covers the full spread across all setpoints; the amber band adds measurement uncertainty (RSS). For the full methodology, see Equipment Monitoring & Drift Analysis.

Figure 9. DC Voltage Source Drift Trend

Figure 10 shows the drift between consecutive calibrations — how much the device changed since it was last calibrated or adjusted. This isolates the device’s instability from corrections applied during calibration.

Figure 10. DC Voltage Source — Drift Between Calibrations